Quote History Quoted:
The meter just had some software to help with looking at wave forms or something. KG can reply to that. I do not think it is good enough to use. Like I said my results were opposite of the ones Kyle expected.That says a lot right there. The string I used was the closest to being the same. I actually was doing some stack development work and was able to collect a lot of data. The meter is useless for that for sure.
View Quote
The information you posted was good, but I am not sure I agree with your conclusion. From what you recorded, I think I am correct iin assuming you were using either A or C weighting and not unweighted (also called Z-weighting). When using A or C, the response and rise time are degraded to that of a 25-or-so microsecond meter. In your string on the 2208, there was a shot-to-shot variation of slightly over 2 dB with similar variation on the LxT.
Had you used unweighted, the difference between the LxT and 2209 would have been on the order of roughly 8 dB rather than the 1.06 that you showed. In that case, I would agree with you without question.
With A or C weighting, the results you obtained with the 2209 are in line with my LD-800B, which has a measured rise time of 13 microseconds. The rise time is inversely related to the highest frequency the meter will measure without distortion, and this frequency defines the sampling rate.
Obviously, the LxT, 800B, and 2209 are not as accurate as the Pulse system, especially unweighted. We had a Pulse demo here at Gemtech a couple of years ago, and compared it to the 800B. Obviously, unweighted, the Pulse read higher. But with C-weighting (closer to correct than A), the difference was minimal. BTW, the 1/4" pressure mic with its 5 microsecond rise time will limit the Pulse's response -- you need the 1/8" pressure mic.
Is the LxT ideal? Definitely not, but it is a lot better than 95% of the meters out there on firearm sounds -- especially when looking at the meters that a number of LE crime labs use. It at least has a peak detector. The industry standard (and permitted in MIL-STD 1474D and before) permit the use of A or C weighting. For these parameters, it is acceptable. The 2209 is considered pretty much the standard of our industry. While the LxT reads slightly over 1 dB low in your example, your round-to-round ammo variation is greater than that. Is averaging 10 rounds adequate? Sure. Would the Std Dev be better at 100 rounds averaged? Possibly. Do you really lose that much averaging only 5 rounds? Not really.
For most purposes, I believe the LxT is adequate, and for the price, it is a better choice than most of what is on the market. If you want really accurate hearing damage risk, bypass the 2208 or 800B and step up to the Pulse system and do
not use weighting.